Testo 320 question
Hello all,
I have a 320 that was recently serviced at Testo. I was doing a combustion test the other day and noticed that the readings changed if I placed the analyzer in a horizontal vs vertical position.
I never noticed this before, but I usually find a good spot for it and leave it. The magnets aren’t the best and it recently slipped down a pipe and the screen landed on a draw off and cracked. So I’m a little more cautious as to where I place it now.
Anyone have any input on this? I haven’t called Testo yet, but I figured some real world input would be nice.
Thank you!
Taki,
  • Zman
    Zman Posts: 7,614
    How much did it change? Have you checked to be sure the lines and filters are clean. I have the same model and will check the next time I use it.
    Albert Einstein
  • STEVEusaPA
    STEVEusaPA Posts: 6,505
    I wonder if in horizontal it's slightly pinching the tubes. I know Bacharach recommends vertical, letting the tubes hang, free of pinching or kinks.
  • ratio
    ratio Posts: 4,329
    I had something similar with a Magnahelic, the readings changed if I moved the gauge. Chased it down to a drop of water in one of the lines—threw everything off.
    I was pulling my hair out for a while!
  • They change quite a bit, over a percent. I swear it didn’t do it before! I called Testo and first the tech said it was out of calibration. But it was calibrated by them less than 6 months ago and has been used a dozen times on residential condensing gas boilers on tune ups.
    I did remove everything and check for water in the lines.
    Now I’m starting to question any adjustments I made to the last few systems I worked on.
    I’ll will update as thing go along. Thanks for the replies!
  • STEVEusaPA
    Do you have access (buddy?) to another analyzer where you can go side by side to check?
  • DZoro
    DZoro Posts: 1,048
    I did have a older 320 do the same thing, I too was concerned. It is now my back up have a new 320. Never put the two side by side for a test. Will do no time for that this winter, maybe later will do a comparison. Just didn't trust it when I seen it do that trick, one needs to trust our test instruments, hence bought new.
    D
  • I do have a access to buddies analyzer, but as mentioned above it’s trying to find the time to get together to do it.
    After another call to Testo I was told to replace the O2 sensor. I ordered one, if I don’t feel confident after that I will have to pull the trigger on a new one. Last time I sent mine out to Testo in Jan they had for a month. That’s too long to be without an analyzer, especially in the winter.