Combustion analyzers
WaHeat
Member Posts: 5
We use the Bacharach combustion analyzers. they seem to need maintenance far too often. is there in house maintenance that you guys know of too keep them analyzing, or maybe a different brand that doesnt require as much upkeep as a Jaguar?
Also, Bacharach's tempoint thermometers are discontinued. Is there another that is up to par?
Also, Bacharach's tempoint thermometers are discontinued. Is there another that is up to par?
0
Comments
-
As soon as i see elevated levels i pull the probe out ! I personally like our Testo...Very easy to change the sensor without having to send it in for service.0
-
I have an UEI and to get it calibrated, I have to sell my first-born to pay for it.0
-
Really? Exaggerating a little?
There was an error rendering this rich post.
0 -
I like my Bacharach but I agree its expensive and often to service . I know service it yourself...I'd rather fix a boiler. Do you'll think Testo is less maintenance AND less expensive? I'm all ears...Mad Dog0
-
I have two 320 which have been discontinued...Great Tool. Not sure,but sometimes you can still find someone may they still have one: https://www.omnicontrols.com/0
-
I've been hearing great things about the Testo 300 lately, but 100% of it has been from the (you guessed it!) guy who sells Testo 300s for a living.Contact John "JohnNY" Cataneo, NYC Master Plumber, Lic 1784
Consulting & Troubleshooting
Heating in NYC or NJ.
Classes5 -
We have a few Testo 330 analyzers. I've never been all that pleased with the sensors or the cost to replace them. Back around the turn of the century (2000) we had three Bacharach analyzers, they were even more finicky.
I feel the current models are overpriced, not any more reliable and very expensive to keep updated. Unfortunately, I'm not sure there is anything better out there. I guess when the united States switches to all electric in a few yeas we'll just throw away our combustion testing equipment???2 -
Never pull out...but if you do...don't shut it off until it completely purges.Derheatmeister said:As soon as i see elevated levels i pull the probe out ! I personally like our Testo...Very easy to change the sensor without having to send it in for service.
There was an error rendering this rich post.
3 -
I change my bacharach sensors myself. Send it in only if there is a problem. But it's been 2 years so I'll probably send it in too TTT.
There was an error rendering this rich post.
0 -
I've been real happy with my Testo 300 that I bought last year for eight bills from Amazon. It was listed as used, but came in the original packaging without any signs of wear. I think it was purchased and then returned without even being turned on.8.33 lbs./gal. x 60 min./hr. x 20°ΔT = 10,000 BTU's/hour
Two btu per sq ft for degree difference for a slab0 -
-
I have been throwing analyzers at my students for 30 years. I don't see them breaking. O2 sensors are the only part that needs changing on a regular basis ( 2 years) but other than
they shouldn't be that expensive to maintain. CO sensors last for 8-10 years. I have never been able to ruin one and I have seen CO near 100,000 ppm.
How often do home alarms that last for 5-10 years get calibrated? I know we want to be safe, but if you are spending money every year for calibration you are wasting money. I have never seen anyone's analyzer off more than 30% and that one just came back from calibration.
If you have two analyzers you can compare their readings. Won't be the same but close. You can also just touch the probe with a flame for a second and if the CO goes near 200 ppm all is good.
Water or moisture is the main problem with analyzers. This does hurt the sensors. However, many times when techs brought me their analyzers and they were wet, I let them sit overnight to dry out and the next day they were fine.
One last thing is the diaphragm pumps that tend to get stuck because of moisture. I have opened up many pumps over the years, loosened the diaphragm and they were fine.
As far as new analyzers, the Sauermann analyzers have some great features including Bluetooth for reading remotely on your I=phone or I-pad.
If you have any more questions, post them here.2 -
@captainco
when your sensor saw 100k ppm CO did you have any issues with it after? I have an E-instruments model with the built in printer (precursor to the current sauerman models i think). One time I maxed out the digital readout for CO so it was well over 10k ppm, it gave me a sensor error for about 3 days. I always just assumed it was the high CO but maybe it was unrelated. CO sensor was just fine, and the next time I had sent it in to replace the O2 sensor the calibration report showed the CO sensor was still fully calibrated.0 -
If you max out your CO sensor you have to keep the pump running until it comes back in range. If you turn off the analyzer and restart it, it could take 2-3 days to recover. The very first Bacharach stopped the pump when the CO was high back in 1985. I figured out a way to short cycle the pump and after 20 minutes got it running again.
I have never had to replace a CO sensor in any analyzer I have used.1 -
Yes...Allways let it Purge..Why would i not pull it out vs. subjecting it to elevated levels ?STEVEusaPA said:
Never pull out...but if you do...don't shut it off until it completely purges.Derheatmeister said:As soon as i see elevated levels i pull the probe out ! I personally like our Testo...Very easy to change the sensor without having to send it in for service.
0 -
That was for anyone who stumbles onto the thread and thinks they should pull it out and shut it off.Derheatmeister said:
Yes...Allways let it Purge..Why would i not pull it out vs. subjecting it to elevated levels ?STEVEusaPA said:
Never pull out...but if you do...don't shut it off until it completely purges.Derheatmeister said:As soon as i see elevated levels i pull the probe out ! I personally like our Testo...Very easy to change the sensor without having to send it in for service.
There was an error rendering this rich post.
0 -
Many times when the CO spikes the O2 reading and flue temperature haven't reached a point where the problem can be diagnosed. If you do pull the probe out what would be the next step without this information?0
-
Hello from Germany ,
I have a Testo 300 in use . I compared it yesterday with a Fieldpiece CAT 85 and have to say that I like the Fieldpiece better .
Testo uses a quarter turn connector plug for the probe hose . The hose is kind of stiff . It happens that the connection is not tight enough and the measurement incorrect.
I also can not get a server connection to send out Service Reports. Tried everything.
Third issue after two years the magnets got so weak I can not attach the unit to a metal surface without sliding down. Not really impressed.
So I will go with the Fieldpiece for sure .
MRU and Dräger are also good but I am not sure if they are sold in the US .
Take care ,
Mario2
Categories
- All Categories
- 86.2K THE MAIN WALL
- 3.1K A-C, Heat Pumps & Refrigeration
- 53 Biomass
- 422 Carbon Monoxide Awareness
- 90 Chimneys & Flues
- 2K Domestic Hot Water
- 5.4K Gas Heating
- 99 Geothermal
- 156 Indoor-Air Quality
- 3.4K Oil Heating
- 63 Pipe Deterioration
- 915 Plumbing
- 6K Radiant Heating
- 381 Solar
- 14.8K Strictly Steam
- 3.3K Thermostats and Controls
- 53 Water Quality
- 41 Industry Classes
- 47 Job Opportunities
- 17 Recall Announcements